| Registry Entry | Optiv Classic 321 Manual | Optiv Classic 321 GL | Optiv Classic 453 | All Other Options | Description |
| Home_SpeedRestrict | TRUE |
TRUE |
TRUE |
TRUE |
Restricts the homing speed. The software displays an error message if the machine exceeds the maximum allowed speed. |
| AxisX_Resolution | -0.00005 |
-0.00005 |
-0.00005 |
-0.00005 |
Defines the X-axis encoder resolution in mm. |
| AxisX_HomeEnabled |
TRUE |
TRUE |
TRUE |
When set to TRUE, the X-axis homing is enabled and required. |
| AxisX_HomeVelocityFactor |
0.25 |
0.05 |
0.25 |
Defines the X-axis homing velocity as a percentage of the maximum speed. The valid values are 0.0 - 1.0 inclusive (0% - 100%). |
| AxisX_HomeFollowingLimitFactor |
0.25 |
0.1 |
0.25 |
Defines the X-axis homing following limit as a percentage of the maximum following limit. The valid values are 0.0 - 1.0 inclusive (0% -100%). |
| AxisY_Resolution | 0.00005 |
0.00005 |
0.00005 |
0.00005 |
Defines the Y-axis encoder resolution in mm. |
| AxisY_HomeEnabled |
TRUE |
TRUE |
TRUE |
When set to TRUE, the Y-axis homing is enabled and required. |
| AxisY_HomeVelocityFactor |
0.25 |
0.05 |
0.25 |
Defines the Y-axis homing velocity as a percentage of the maximum speed. The valid values are 0.0 - 1.0 inclusive (0% -100%). |
| AxisY_HomeFollowingLimitFactor |
0.25 |
0.1 |
0.25 |
Defines the Y-axis homing following limit as a percentage of the maximum following limit. The valid values are 0.0 - 1.0 inclusive (0% -100%). |
| AxisZ_Resolution | -0.00005 |
-0.00005 |
-0.00005 |
-0.00005 |
Defines the Z-axis encoder resolution in mm. |
| AxisZ_HomeEnabled |
TRUE |
TRUE |
TRUE |
When set to TRUE, the Z-axis homing is enabled and required. |
| AxisZ_HomeVelocityFactor |
0.25 |
0.05 |
0.25 |
Defines the Z-axis homing velocity as a percentage of the maximum speed. The valid values are 0.0 - 1.0 inclusive (0% -100%). |
| AxisZ_HomeFollowingLimitFactor |
0.25 |
0.1 |
0.25 |
Defines the Z-axis homing following limit as a percentage of the maximum following limit. The valid values are 0.0 - 1.0 inclusive (0% -100%). |
| AxisQ_Resolution |
0.006 |
0.006 |
0.006 |
Defines the rotary Q-axis encoder resolution in degrees. |
| AxisQ_HomeEnabled |
FALSE |
FALSE |
FALSE |
When set to TRUE, the rotary Q-axis homing is enabled and required. |
| Zoom_Resolution | -0.001 |
-0.001 |
-0.001 |
-0.001 |
Defines the zoom axis encoder resolution. |
| Zoom_Velocity | 6.0 |
Defines the zoom axis velocity. |
| Zoom_Acceleration | 45.0 |
Defines the zoom axis acceleration. |
| Zoom_Jerk | 0.0 |
Defines the zoom axis tuning parameter Jerk. |
| Zoom_PFactor | 50.0 |
Defines the zoom axis tuning parameter Proportional Gain Factor. |
| Zoom_DFactor | 0.0 |
Defines the zoom axis tuning parameter Derivative Gain Factor. |
| Zoom_IFactor | 0.0 |
Defines the Zoom axis tuning parameter Integral Gain Factor. |
| Zoom_ILimit | 0.0 |
Defines the zoom axis tuning parameter Integral Limit. |
| Zoom_ContThreshold | 10.0 |
Defines the zoom axis tuning parameter Continuation Threshold. |
| Zoom_FollowingLimit | 0.05 |
Defines the zoom axis tuning parameter Following Error Limit. |
| Zoom_BacklashCorrection | 500 |
500 |
500 |
500 |
Defines the zoom overshoot distance to counter Backlash (encoder counts) Correction. |
| Illumination_SettleTime | 0.1 |
0.1 |
0.1 |
0.1 |
Defines the delay in seconds for the illumination to settle after a change. |
| Stage_SettleTime |
0.1 |
0.1 |
0.1 |
Defines the delay in seconds for the stage movement to settle after the software detects a motion. |
| Zoom_SettleTime | 0.1 |
0.1 |
0.1 |
0.1 |
Defines the delay in seconds for the optics to settle after a magnification change occurs. |
| Pendant_EStopInhibitsAmp |
TRUE |
TRUE |
TRUE |
When set to TRUE, the emergency stop (ESTOP) disables the drive amps. |
| Pendant_FineSpeedFactor |
1.0 |
1.0 |
1.0 |
Defines the fine speed as a percentage of the maximum speed. The valid values are 0.0 - 1.0 inclusive (0% - 100%). |
| Pendant_X_Reversed |
FALSE |
When set to TRUE, the X-axis pendant motion is reversed. |
| Pendant_X_Deadband |
0.2 |
Defines the X-axis pendant motion deadband at the center position in percent. |
| Pendant_X_Speed |
1.0 |
Defines the X-axis pendant movement speed. |
| Pendant_X_FineSpeed |
0.1 |
Defines the X-axis pendant fine movement speed. |
| Pendant_X_Exponential |
0.5 |
Defines the X-axis pendant motion curve. |
| Pendant_Y_Reversed |
TRUE |
When set to TRUE, the Y-axis pendant motion is reversed. |
| Pendant_Y_Deadband |
0.2 |
Defines the Y-axis pendant motion deadband at the center position in percent. |
| Pendant_Y_Speed |
1.0 |
Defines the Y-axis pendant movement speed. |
| Pendant_Y_FineSpeed |
0.1 |
Defines the Y-axis pendant fine movement speed. |
| Pendant_Y_Exponential |
0.5 |
Defines the Y-axis pendant motion curve. |
| Pendant_Z_Reversed |
FALSE |
When set to TRUE, the Z-axis pendant motion is reversed. |
| Pendant_Z_Deadband |
0.1 |
Defines the Z-axis pendant motion deadband at the center position in percent. |
| Pendant_Z_Speed |
1.0 |
Defines the Z-axis pendant movement speed. |
| Pendant_Z_FineSpeed |
0.1 |
Defines the Z-axis pendant fine movement speed. |
| Pendant_Z_Exponential |
0.5 |
Defines the Z-axis pendant motion curve. |
| Pendant_Q_Reversed |
FALSE |
When set to TRUE, the rotary Q-axis pendant motion is reversed. |
| Pendant_Q_Deadband |
0.1 |
Defines the rotary Q-axis pendant motion deadband at the center position in percent. |
| Pendant_Q_Speed |
1.0 |
Defines the rotary Q-axis pendant movement speed. |
| Pendant_Q_FineSpeed |
0.1 |
Defines the rotary Q-axis pendant fine movement speed. |
| Pendant_Q_Exponential |
0.5 |
Defines the rotary Q-axis pendant motion curve. |
| Probe_DebounceTime |
0.05 |
0.05 |
0.05 |
Defines the time in seconds that the touch probe must be triggered to detect a hit. |
| Probe_DirectionThreshold |
0.01 |
0.01 |
0.01 |
Defines the distance in mm that the touch probe must move to determine the hit vector. |
| Probe_DisableIfNotActive |
TRUE |
TRUE |
TRUE |
When set to TRUE, the software disables the touch probe if it is not the active probe. |
| Probe_UsePI200 |
FALSE |
FALSE |
FALSE |
When set to TRUE, the software uses the Renishaw PI200/TP200 probe. |
