Home > Available Help Files > PC-DMIS CMM > Working with Measurement Strategies > Using Non-Adaptive Scanning Strategies > Cylinder Centering Thread Scan Strategy
The Cylinder Centering Thread Scan strategy for the Cylinder Auto feature performs a thread scan by maintaining the probe centered within the thread. When you use this strategy, the diameter of the probe tip must be larger than the size of the valleys in-between the threads to prevent probe shanking.
This strategy is supported only on B4 and B5 Leitz controllers.
The following properties are available:
Sample Cylinder Centering Thread Scan properties
Point Density
Type or select the number of readings to take per unit of measurement during the scan.
Scan Speed
Type or select the scan speed (mm/sec).
Find Nadir
Select this check box to take two hits at slightly different points on the thread to determine the best place to start the scan. The software uses the point that is deepest into the thread.
Filtering Area
Filter Type - Select the filter type:
None - This option does not apply any filter type to the scan data set.
Gaussian - This option applies a Gaussian cylindrical filter to the scan data set.
UPR - This option defines the undulations per revolution. The default is 50. This value only applies to cylinders and circles. The software hides this option if you select None in the Filter Type list.
