Example of Multiple Probe References
Suppose you have these three probes configurations:
PROBE_01
PROBE_02
PROBE_03
QuickConnect AutoJoint
TP2
TP20
TP20
3mmx10mm stylus
2mmx10mm stylus
4mmx20mm stylus
The ACR1 will switch between the TP2 and the TP20 probes. The MCR20 will switch between the styli associated with the TP20 probe system(s).
A typical port definition might look like this:
ACR1
MCR20
Port 1
Port 2
Probe_01
Probe_02
Probe_03
In operation, suppose the system is using PROBE_01 and it needs to switch to PROBE_02. The system would:
Suppose after measuring some features, the system then needs to use PROBE_03. The system would:
Now, suppose the system needs to switch from PROBE_03 (the TP20 with the 4mmx20mm stylus) back to Probe_01 (the TP2 with the 3mmx10mm stylus). The system would:
Note that in this example, you only need one TP20 probe body. You use the MCR20 to switch between two different probe stylus assemblies for varied measurement requirements.