ProbeQualVisionFocusMaxDeviationPercentage
The value for this entry is the percentage of the set frame rate and defines the maximum permitted deviation of the set frame rate from the frame rate determined during calibration. If this limit value is exceeded, PC-DMIS displays a warning message.
For example, if a frame rate of 25 is communicated to the frame grabber (set frame rate), and you set this limit value to 15, the maximum deviation of the frame rate determined during calibration is a maximum of 3.75. In other words, if the frame rate is less than 21.25, PC-DMIS issues a warning message.
Entry Name: ProbeQualVisionFocusMaxDeviationPercentage
Entry Type: Real Number to specify the percentage of the set frame rate. The default value is 15.