Home > Available Help Files > Machine Interface Installation Manual (MIIM) > CMM Supported Interfaces > Leitz-Common Firmware Interface > Common Firmware Errors
Current Help File: PC-DMIS MIIM (View Core Help)
You can diagnose errors and failures that may occur during the operation of the Common Firmware as follows:
If an error occurs during the execution of a command, review the error message that the controller immediately sends.
If the interface software is active, review the error message that appears.
The right-hand columns in the tables below list the error messages that may appear.
The controller's firmware stores the last error condition until you reapply power. You can then analyze the condition. To analyze it, you need to connect to the controller through the HyperTerminal application and STATUS command, or use the TESTSOFT program and function 1 SYSTEM STATUS. A series of informational messages appears and displays an "Error" field that contains a STATUS label with the type of error. The left-hand columns in the tables below list the STATUS labels that may appear.
STATUS Label |
Error Message |
Cause |
excess_lag : |
ERROR MOTOR_RE,S%1,TR_FG002,Excess Lag Error |
The firmware detected that the position (the real axis position with respect to the position set) exceeded the maximum limit. This error may be due to one of the following:
STATUS Label |
Error Message |
Cause |
x_oversp_error : |
ERROR X_OVSPD ,S%1 ,TR_FG16e, Overspeed Error |
The firmware detected that the speed (the speed obtained from the counts with respect to the speed command sent to the D/A) exceeded the maximum limit. This error may be due to one of the following:
y_oversp_error : |
ERROR Y_OVSPD ,S%1 ,TR_FG16e, Overspeed Error |
z_oversp_error : |
ERROR Z_OVSPD ,S%1 ,TR_FG16e, Overspeed Error |
w_oversp_error : |
ERROR W_OVSPD ,S%1 ,TR_FG16e, Overspeed Error |
a_oversp_error : |
ERROR A_OVSPD ,S%1 ,TR_FG16e, Overspeed Error |
b_oversp_error : |
ERROR B_OVSPD ,S%1 ,TR_FG16e, Overspeed Error |
STATUS Label |
Error Message |
Cause |
excess_prb_dfl : |
ERROR MOTOR_RE,S%1,TR_FG003,Excess Probe Deflection |
Excessive probe deflection occurred. |
STATUS Label |
Error Message |
Cause |
x_pos_hw_limit : |
ERROR MOTOR_X+,S%1,TR_FG004,Plus Hardware Limit Error |
The firmware detected that the hardware travel limit is not in the expected position, or it is in an unexpected position. The position transducer may be defective, or the travel limit position for the M_HWLIM and P_HWLIM parameters of the CONSTANT is incorrect. |
y_pos_hw_limit : |
ERROR MOTOR_Y+,S%1,TR_FG004,Plus Hardware Limit Error |
z_pos_hw_limit : |
ERROR MOTOR_Z+,S%1,TR_FG004,Plus Hardware Limit Error |
w_pos_hw_limit : |
ERROR MOTOR_W+,S%1,TR_FG004,Plus Hardware Limit Error |
a_pos_hw_limit : |
ERROR MOTOR_A+,S%1,TR_FG004,Plus Hardware Limit Error |
b_pos_hw_limit : |
ERROR MOTOR_B+,S%1,TR_FG004,Plus Hardware Limit Error |
x_neg_hw_limit : |
ERROR MOTOR_X-,S%1,TR_FG005,Minus Hardware Limit Error |
y_neg_hw_limit : |
ERROR MOTOR_Y-,S%1,TR_FG005,Minus Hardware Limit Error |
z_neg_hw_limit : |
ERROR MOTOR_Z-,S%1,TR_FG005,Minus Hardware Limit Error |
w_neg_hw_limit : |
ERROR MOTOR_W-,S%1,TR_FG005,Minus Hardware Limit Error |
a_neg_hw_limit : |
ERROR MOTOR_A-,S%1,TR_FG005,Minus Hardware Limit Error |
b_neg_hw_limit : |
STATUS Label |
Error Message |
Cause |
x_both_hw_limit : |
ERROR MOTOR_X?,S%2,TR_FG004,Plus Hardware Limit Error |
Both hardware travel limits are active. |
y_both_hw_limit : |
ERROR MOTOR_Y?,S%2,TR_FG004,Plus Hardware Limit Error |
z_both_hw_limit : |
ERROR MOTOR_Z?,S%2,TR_FG004,Plus Hardware Limit Error |
w_both_hw_limit : |
ERROR MOTOR_W?,S%2,TR_FG004,Plus Hardware Limit Error |
a_both_hw_limit : |
ERROR MOTOR_A?,S%2,TR_FG004,Plus Hardware Limit Error |
b_both_hw_limit : |
ERROR MOTOR_B?,S%2,TR_FG004,Plus Hardware Limit Error |
STATUS Label |
Error Message |
Cause |
x_pos_sw_limit : |
ERROR MOTOR_X+,S%3,TR_FG030,Plus Software Limit Error |
The software travel limit was reached. |
y_pos_sw_limit : |
ERROR MOTOR_Y+,S%3,TR_FG030,Plus Software Limit Error |
z_pos_sw_limit : |
ERROR MOTOR_Z+,S%3,TR_FG030,Plus Software Limit Error |
w_pos_sw_limit : |
ERROR MOTOR_W+,S%3,TR_FG030,Plus Software Limit Error |
a_pos_sw_limit : |
ERROR MOTOR_A+,S%3,TR_FG030,Plus Software Limit Error |
b_pos_sw_limit : |
ERROR MOTOR_B+,S%3,TR_FG030,Plus Software Limit Error |
x_neg_sw_limit : |
ERROR MOTOR_X-,S%3,TR_FG031,Minus Software Limit Error |
y_neg_sw_limit : |
ERROR MOTOR_Y-,S%3,TR_FG031,Minus Software Limit Error |
z_neg_sw_limit : |
ERROR MOTOR_Z-,S%3,TR_FG031,Minus Software Limit Error |
w_neg_sw_limit : |
ERROR MOTOR_W-,S%3,TR_FG031,Minus Software Limit Error |
a_neg_sw_limit : |
ERROR MOTOR_A-,S%3,TR_FG031,Minus Software Limit Error |
b_neg_sw_limit : |
ERROR MOTOR_B-,S%3,TR_FG031,Minus Software Limit Error |
STATUS Label |
Error Message |
Cause |
e_stop_p_b : |
ERROR MOTOR_RE,S%1,TR_FG013,E-Stop Push-Button Pressed |
The emergency (mushroom-shaped) pushbutton was pressed. |
STATUS Label |
Error Message |
Cause |
PH9_error : |
ERROR MOTOR_RE,S%1,TR_FG014,Wrist Error |
An error on the PH9 or PH10 probe head occurred. |
STATUS Label |
Error Message |
Cause |
Wrist_error : |
An error on the PHS or CW43 probe head occurred. |
STATUS Label |
Error Message |
Cause |
e_stop_relay : |
ERROR MOTOR_RE,S%1,TR_FG015,E-Stop Relay Open |
The motor relay is open. |
STATUS Label |
Error Message |
Cause |
home_error : |
ERROR MOTOR_RE,S%1,TR_FG018,Homing Cycle Error |
An error in the reset cycle occurred. |
STATUS Label |
Error Message |
Cause |
air_error : |
ERROR MOTOR_RE,S%1,TR_FG023,Machine Air Pressure or Lube Fault |
Air is missing. |
STATUS Label |
Error Message |
Cause |
Mach Temp fault : |
ERROR MOTOR_RE,S%1,TR_FG024,Machine or Controller Temperature Fault |
The temperature in the cabinet is too high. |
STATUS Label |
Error Message |
Cause |
Illegal Unlock : |
ERROR MOTOR_RE,S%1,TR_FG025,Illegal Unlock Condition |
The "Unlock" input condition is not allowed. |
STATUS Label |
Error Message |
Cause |
No Constant File : |
ERROR MOTOR_RE,S%1,TR_FG020,CONSTANT.ASC file read Error |
The CONSTANT file is missing. |
STATUS Label |
Error Message |
Cause |
Drive Fault: |
ERROR MOTOR_RE,S%1,TR_SERVO,Servo Amplifier Not Ready! |
The servodrivers are not ready. |
STATUS Label |
Error Message |
Cause |
InPosition Error : |
ERROR MOTOR_RE,S%1,TR_FG028,In-Position error time-out |
The axes do not reach the position set on time. This error may be due to one of the following:
STATUS Label |
Error Message |
Cause |
Host Stop Demand : |
ERROR MOTOR_RE,S%1,TR_FG029,Stop Requested by Host |
The host sent an abort request. |
STATUS Label |
Error Message |
Cause |
Jogbox Fault : |
ERROR MOTOR_RE,S%1,TR_FG032,Jogbox Fault |
A jog box error occurred. |
STATUS Label |
Error Message |
Cause |
Ram_Collision : |
ERROR B3C_Ram,S%1,TR_BUE04,Ram Collision! |
The anti-crash device operates. |
STATUS Label |
Error Message |
Cause |
X_Counter_Error : |
ERROR X_CNTR,S%1,TR_FG100,X Counter Error |
The error of the digital count signals; the switching gap of the count signals is too narrow. |
Y_Counter_Error : |
ERROR Y_CNTR,S%1,TR_FG101,Y Counter Error |
Z_Counter_Error : |
ERROR Z_CNTR,S%1,TR_FG102,Z Counter Error |
W_Counter_Error : |
ERROR W_CNTR,S%1,TR_FG103,W Counter Error |
A_Counter_Error : |
ERROR A_CNTR,S%1,TR_FG104,A Counter Error |
B_Counter_Error : |
ERROR B_CNTR,S%1,TR_FG105,B Counter Error |
STATUS Label |
Error Message |
Cause |
X_Traser_Error : |
ERROR X_Traser,S%1,TR_FG106,X Traser Error |
The transducers signaled an error. |
Y_Traser_Error : |
ERROR Y_Traser,S%1,TR_FG107,Y Traser Error |
Z_Traser_Error : |
ERROR Z_Traser,S%1,TR_FG108,Z Traser Error |
W_Traser_Error : |
ERROR W_Traser,S%1,TR_FG109,W Traser Error |
A_Traser_Error : |
ERROR A_Traser,S%1,TR_FG10a,A Traser Error |
B_Traser_Error : |
ERROR B_Traser,S%1,TR_FG10b,B Traser Error |
STATUS Label |
Error Message |
Cause |
X_Flet_Error : |
ERROR X_FLET,S%1,TR_FG110,X Flet Error |
The digital count signal is missing. |
Y_Flet_Error : |
ERROR Y_FLET,S%1,TR_FG111,Y Flet Error |
Z_Flet_Error : |
ERROR Z_FLET,S%1,TR_FG112,Z Flet Error |
W_Flet_Error : |
ERROR W_FLET,S%1,TR_FG113,W Flet Error |
A_Flet_Error : |
ERROR A_FLET,S%1,TR_FG114,A Flet Error |
B_Flet_Error : |
ERROR B_FLRT,S%1,TR_FG115,B Flet Error |
STATUS Label |
Error Message |
Cause |
X_AmplitudeError : |
ERROR X_Amplit,S%1,TR_FG116,X Amplitude Error |
The width of the analog count signal is wrong. |
Y_AmplitudeError : |
ERROR Y_Amplit,S%1,TR_FG117,Y Amplitude Error |
Z_AmplitudeError : |
ERROR Z_Amplit,S%1,TR_FG118,Z Amplitude Error |
W_AmplitudeError : |
ERROR W_Amplit,S%1,TR_FG119,W Amplitude Error |
A_AmplitudeError : |
ERROR A_Amplit,S%1,TR_FG11a,A Amplitude Error |
B_AmplitudeError : |
ERROR B_Amplit,S%1,TR_FG11b,B Amplitude Error |
STATUS Label |
Error Message |
Cause |
X_Phase_Error : |
ERROR X_Phase ,S%1,TR_FG120,X Phase Error |
Count error. The digital count and the value of the analog signal do not match. |
Y_Phase_Error : |
ERROR Y_Phase ,S%1,TR_FG121,Y Phase Error |
Z_Phase_Error : |
ERROR Z_Phase ,S%1,TR_FG122,Z Phase Error |
W_Phase_Error : |
ERROR W_Phase ,S%1,TR_FG123,W Phase Error |
A_Phase_Error : |
ERROR A_Phase ,S%1,TR_FG124,A Phase Error |
B_Phase_Error : |
ERROR B_Phase ,S%1,TR_FG125,B Phase Error |
STATUS Label |
Error Message |
Cause |
BadModeSwitch : |
ERROR BadSwtch,S%1,TR_FG130,Bad Mode Switch Selection |
The MAN/AUTO selector switch is in an undefined position. |
STATUS Label |
Error Message |
Cause |
ModeSelFailure : |
ERROR ModFail ,S%1,TR_FG131,Mode Selection Fatal Failure |
The MAN/AUTO switching was done while the axes are moving. |
STATUS Label |
Error Message |
Cause |
JogKeybdFailure : |
ERROR JogKeybd,S%1,TR_FG132,JogBox Key Fatal Failure |
A jog box keyboard error occurred. |
STATUS Label |
Error Message |
Cause |
D_Drive Err_set : |
ERROR DualDriv,S%1,TR_FG16f,Dual Drive Settings Error |
An error in the definition of the Dual Drive parameters in the CONSTANT.ASC file occurred (the type is not recognized, or the slave axis is incompatible). |
STATUS Label |
Error Message |
Cause |
D_Drive Err_LVDT : |
ERROR DualDriv,S%1,TR_FG170,Dual Drive LVDT Error |
On an axis of the Dual Drive type, this error signals that the hardware detected an excessive difference in scale information between LVDT’s. |
STATUS Label |
Error Message |
Cause |
DualDrive : |
ERROR DualDriv,S%1,TR_FG133,Dual Drive Disalignment Error |
On an axis of the Dual Drive type, this error signals that the firmware detected a misalignment between the master and slave axes. |
STATUS Label |
Error Message |
Cause |
Power_Failure : |
ERROR PowrFail,S%1,TR_FG134,Power Check Fatal Failure |
The power relay status disagrees with the command. |
STATUS Label |
Error Message |
Cause |
X_Overcurrent : |
ERROR XOvrCurr,S%1,TR_FG140,X Overcurrent Error |
A quick-acting circuit detected an anomalous input of the final stages of the servodrivers. This error may be due to a short circuit, such as on the motor cable. |
Y_Overcurrent : |
ERROR YOvrCurr,S%1,TR_FG141,Y Overcurrent Error |
Z_Overcurrent : |
ERROR ZOvrCurr,S%1,TR_FG142,Z Overcurrent Error |
W_Overcurrent : |
ERROR WOvrCurr,S%1,TR_FG143,W Overcurrent Error |
A_Overcurrent : |
ERROR AOvrCurr,S%1,TR_FG144,A Overcurrent Error |
B_Overcurrent : |
ERROR BOvrCurr,S%1,TR_FG145,B Overcurrent Error |
STATUS Label |
Error Message |
Cause |
X_Tacho_Fault : |
ERROR X_Tacho,S%1,TR_FG146,X Tacho Error |
During operation in Manual mode, a circuit on the servodriver detected that the tacho (speed) exceeded the maximum level. |
Y_Tacho_Fault : |
ERROR Y_Tacho,S%1,TR_FG147,Y Tacho Error |
Z_Tacho_Fault : |
ERROR Z_Tacho,S%1,TR_FG148,Z Tacho Error |
W_Tacho_Fault : |
ERROR W_Tacho,S%1,TR_FG149,W Tacho Error |
A_Tacho_Fault : |
ERROR A_Tacho,S%1,TR_FG14a,A Tacho Error |
B_Tacho_Fault : |
ERROR B_Tacho,S%1,TR_FG14b,B Tacho Error |
STATUS Label |
Error Message |
Cause |
X_IxT : |
ERROR X_IxT,S%1,TR_FG150,X IxT Error |
A circuit on the servodriver detected saturation (that is, the peak level) of the motor current for a time that is longer than allowed. This error may be due to one of the following:
Y_IxT : |
ERROR Y_IxT,S%1,TR_FG151,Y IxT Error |
Z_IxT : |
ERROR Z_IxT,S%1,TR_FG152,Z IxT Error |
W_IxT : |
ERROR W_IxT,S%1,TR_FG153,W IxT Error |
A_IxT : |
ERROR A_IxT,S%1,TR_FG154,A IxT Error |
B_IxT : |
ERROR B_IxT,S%1,TR_FG155,B IxT Error |
STATUS Label |
Error Message |
Cause |
PWM_CnctFailure : |
ERROR PWM_Cnt,S%1,TR_FG160,PWM Connector Failure |
The connection of the analog connectors of the PWM card failed. |
STATUS Label |
Error Message |
Cause |
PWM_PowerFail : |
ERROR PWM_Pwr,S%1,TR_FG161,PWM Power Failure |
A circuit on the servodriver detected errors on some internal supply voltages. |
STATUS Label |
Error Message |
Cause |
PWM_VpHihg : |
ERROR PWMVpHig,S%1,TR_FG162,PWM VpHigh |
The voltage to the servodrivers is too high. |
STATUS Label |
Error Message |
Cause |
PWM_VpLow : |
ERROR PWMVpLow,S%1,TR_FG163,PWM VpLow |
The voltage to the servodrivers is too low. |
STATUS Label |
Error Message |
Cause |
PWM_OverTemp : |
ERROR PWMOvTmp,S%1,TR_FG164,PWM Over Temperature |
The servodriver temperature is too high. |
STATUS Label |
Error Message |
Cause |
PWM_Refer : |
ERROR PWMRef,S%1,TR_FG165,PWM Refer Error |
A circuit on the servodriver detected that the reference voltage 5 Volt is incorrect. |
STATUS Label |
Error Message |
Cause |
PWM_NoMotor : |
ERROR PWMNoMot,S%1,TR_FG166,PWM No Motor Error |
A circuit on the servodriver detected that the tacho connector is not inserted. |
STATUS Label |
Error Message |
Cause |
PWM_NoRpwm : |
ERROR PWMNoRpw,S%1,TR_FG167,PWM no RPWM Error |
The READY signals from servodrivers are missing. This error is a general error signaling that the servodriver detected without any further explanation. |
STATUS Label |
Error Message |
Cause |
X_RefMark_Error : |
ERROR X_RefMrk,S%1,TR_FG168,X Reference Mark Synchro Error |
The firmware did not detect a zero notch where it should be, or it detected a notch in an unexpected position. The position transducer may be faulty, or the ZER_PER or ZER_PREC parameters of the CONSTANT are incorrect. |
Y_RefMark_Error : |
ERROR Y_RefMrk,S%1,TR_FG169,Y Reference Mark Synchro Error |
Z_RefMark_Error : |
ERROR Z_RefMrk,S%1,TR_FG16a,Z Reference Mark Synchro Error |
W_RefMark_Error : |
ERROR W_RefMrk,S%1,TR_FG16b,W Reference Mark Synchro Error |
A_RefMark_Error : |
ERROR A_RefMrk,S%1,TR_FG16c,A Reference Mark Synchro Error |
B_RefMark_Error : |
ERROR B_RefMrk,S%1,TR_FG16d,B Reference Mark Synchro Error |
STATUS Label |
Error Message |
Cause |
SafSys_Violation : |
ERROR MOTOR_RE,S%1,TR_SAF00,Safety System Violation |
The guards are open. |
STATUS Label |
Error Message |
Cause |
Illegal_cha_mode : |
ERROR MOTOR_RE,S%1,TR_SAF01,Illegal Change Mode [Single/Double Arm] |
A fault on the single arm/dual arm switching occurred. |
STATUS Label |
Error Message |
Cause |
Eme_from_f2 : |
ERROR MOTOR_RE,S%1,TR_SAF02,Error Induced from the other Arm! |
The other arm caused an emergency. |
STATUS Label |
Error Message |
Cause |
HoldPanel_NotOk : |
ERROR MOTOR_RE,S%1,TR_SAF03,Hold button on Panel not Ok! |
The panel's "Hold" pushbutton was mistakenly activated. |
STATUS Label |
Error Message |
Cause |
StartPanel_NotOk : |
ERROR MOTOR_RE,S%1,TR_SAF04,Start button on Panel not Ok! |
The panel's "Run" pushbutton was mistakenly activated. |
STATUS Label |
Error Message |
Cause |
MotonPanel_NotOk : |
ERROR MOTOR_RE,S%1,TR_SAF05,Motor On button on Panel not Ok! |
The panel's "Start" pushbutton was mistakenly activated. |
STATUS Label |
Error Message |
Cause |
PWM_PwS2Rack : |
ERROR PwS2Rack,S%1,TR_FG171,Power Supply on rack 2 |
An error on the motor power supply of the second rack occurred. |
STATUS Label |
Error Message |
Cause |
PWM_BrakeResi : |
ERROR BrkResi,S%1,TR_FG172,Motor Braking Resistor |
A braking resistance failure occurred. |
STATUS Label |
Error Message |
Cause |
PWM_OvcAllAx : |
ERROR OvcAllAx,S%1,TR_FG173,Overcurrent in 1 of the 3 axes |
This error applies only to the PWM3.28 card. This error is signaled when one of the three axis servodrivers is faulty. It signals that a quick-acting circuit has detected a wrong input of the final stages of the servodrivers. It may be due to a short circuit, such as on the motor cable. |
STATUS Label |
Error Message |
Cause |
FDG_CYC : |
ERROR MOTOR_RE,S%1,TR_SAF07,FDG cycle error |
The ENABLE CYCLE device of the FDG indicates that the part did not load correctly while an automatic measuring cycle was starting; this error never generates if the modal selector is in the "MAN" position. This diagnostic test is only enabled if the constant HAVE_FDG parameter is set to 1. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR MOTOR_RE,S%1,TR_FG001,E-Stop - waiting for machine start |
The machine is in an E-Stop condition. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR CYGPRCH,S%1,TR_NZERO,Machine not zeroed! |
The machine did not complete the axes zero presetting procedure. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR LTC86,S%1,TR_FG020,CONSTANT.ASC file read Error |
An error in the CONSTANT file reading occurred. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR SET_PAR,S%1,TR_NOCON,CONSTANT.ASC missing, must be downloaded |
The CONSTANT file does not exist. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR LTC86,S%1,TR_TSENS,Temperature Sensor defect, Temp Corr off! |
An error on a temperature sensor occurred, and temperature compensation was deactivated. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR LTC86,S%1,TR_EXE00,invalid Command |
A command is invalid. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR LTC86,S%1,TR_EXE01,invalid Parameter! |
A parameter is invalid. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR LTC86,S%1,TR_NOTIM,Command not implemented! |
The command does not exist. |
STATUS Label |
Error Message |
Cause |
no_error : | ERROR CNVSTREA,S%1,TR_LIB04,Conversion Error String -> Number! | An error in the string-number conversion occurred. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR PULT_PAR,S%1,TR_LIB01,Syntax Error! |
A command syntax error occurred. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR PULT_PAR,S%1,TR_LIB06,too many Parameters! |
The number of parameters is too high. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR ABORT,S%1,TR_MOV01,Movement Interrupted! |
The movement (no fly) was interrupted due to probe deflection. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR FLY,S%1,TR_MOV01,Movement Interupted! |
The movement (fly) was interrupted due to probe deflection. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR <X_LIMIT>,S%1,TR_OUTOL,Machine Parameter out of tolerance! |
The specified stroke is beyond the limits of the axis. |
no_error : |
ERROR <Y_LIMIT>,S%1,TR_OUTOL,Machine Parameter out of tolerance! |
no_error : |
ERROR <Z_LIMIT>,S%1,TR_OUTOL,Machine Parameter out of tolerance! |
no_error : |
ERROR <A_LIMIT>,S%1,TR_OUTOL,Machine Parameter out of tolerance! |
no_error : |
ERROR <B_LIMIT>,S%1,TR_OUTOL,Machine Parameter out of tolerance! |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR PROBE,S%1,TR_PRB01,no Probing within Probelimit! |
There is no measurement within the space. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR GET_PARA,S%1,TR_MINM2,Machine Parameter larger than permitted |
The value of a specified parameter is greater than the maximum limit. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR GET_PARA,S%1,TR_MINM1,Machine Parameter smaller than permitted |
The value of a specified parameter is smaller than the minimum limit. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR GET_PARA,S%1,TR_SCA01,Too many scan control points (max. 8)! |
A specified number of scanning reference points is too high (the maximum is 8). |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR FSCAN,S%1,TR_CEN02,Workpiece contact lost! |
Contact was lost during scanning. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR TRANSFOR,S%1,TR_SCA03,Scanning Direction ambiguous! |
The scanning direction is undefined. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR SET_PAR,S%1,TR_SPMTX,Prb deflection(s) too large for sp6_mtrx |
There is excessive SP600 deflection. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR WRIST,S%1,TR_PH902,Illegal Wrist Angle Commanded |
The wrist position is not allowed. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR WRIST,S%1,TR_PH901,Wrist Not in Configuration |
A wrist is not present in the configuration. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR LOAD_EEP,W%1,TR_LOADK,No machine correction data available! |
Geometric compensation data is unavailable. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR READ_LIN,W%1,TR_TRMI1,RS232 ERROR : BUFFER OVERFLOW |
An error of the serial line occurred. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR DEACOMP,S%1,TR_DEACT,Wrong temp for DEA Activ, Temp Corr off! |
One or more pairs of thermal sensors that Active compensation uses indicates a difference in temperature that is incompatible with the limits that the algorithm set. You should use the function "Board sub-menu" -> "Test Temp Board" -> "Display DEA Activ" of TestSoft to have information on which sensors caused the problem. |
STATUS Label |
Error Message |
Cause |
no_error : |
ERROR DEACOMP,S%1,TR_NOTER,Sensor bank not termin., Temp Corr off! |
On one of the banks of thermal sensors that Active compensation uses, an error occurred in the termination resistors that close the unused channels. You should use the function "Board sub-menu" -> "Test Temp Board" -> "Display Data" of TestSoft to have information on which termination resistors caused the problem. |