Setting Entry | Optiv Classic 321 Manual | Optiv Classic 321 GL | Optiv Classic 453 | All Other Options | Description |
Home_SpeedRestrict | True |
True |
True |
True |
Restricts the homing speed. The software displays an error message if the machine exceeds the maximum allowed speed. |
AxisX_Resolution | -0.00005 |
-0.00005 |
-0.00005 |
-0.00005 |
Defines the X-axis encoder resolution in mm. |
AxisX_HomeEnabled |
True |
True |
True |
When set to True, the X-axis homing is enabled and required. |
AxisX_HomeVelocityFactor |
0.25 |
0.05 |
0.25 |
Defines the X-axis homing velocity as a percentage of the maximum speed. The valid values are 0.0 - 1.0 inclusive (0% - 100%). |
AxisX_HomeFollowingLimitFactor |
0.25 |
0.1 |
0.25 |
Defines the X-axis homing following limit as a percentage of the maximum following limit. The valid values are 0.0 - 1.0 inclusive (0% -100%). |
AxisY_Resolution | 0.00005 |
0.00005 |
0.00005 |
0.00005 |
Defines the Y-axis encoder resolution in mm. |
AxisY_HomeEnabled |
True |
True |
True |
When set to True, the Y-axis homing is enabled and required. |
AxisY_HomeVelocityFactor |
0.25 |
0.05 |
0.25 |
Defines the Y-axis homing velocity as a percentage of the maximum speed. The valid values are 0.0 - 1.0 inclusive (0% -100%). |
AxisY_HomeFollowingLimitFactor |
0.25 |
0.1 |
0.25 |
Defines the Y-axis homing following limit as a percentage of the maximum following limit. The valid values are 0.0 - 1.0 inclusive (0% -100%). |
AxisZ_Resolution | -0.00005 |
-0.00005 |
-0.00005 |
-0.00005 |
Defines the Z-axis encoder resolution in mm. |
AxisZ_HomeEnabled |
True |
True |
True |
When set to True, the Z-axis homing is enabled and required. |
AxisZ_HomeVelocityFactor |
0.25 |
0.05 |
0.25 |
Defines the Z-axis homing velocity as a percentage of the maximum speed. The valid values are 0.0 - 1.0 inclusive (0% -100%). |
AxisZ_HomeFollowingLimitFactor |
0.25 |
0.1 |
0.25 |
Defines the Z-axis homing following limit as a percentage of the maximum following limit. The valid values are 0.0 - 1.0 inclusive (0% -100%). |
AxisQ_Resolution |
0.006 |
0.006 |
0.006 |
Defines the rotary Q-axis encoder resolution in degrees. |
AxisQ_HomeEnabled |
False |
False |
False |
When set to True, the rotary Q-axis homing is enabled and required. |
Zoom_Resolution | -0.001 |
-0.001 |
-0.001 |
-0.001 |
Defines the zoom axis encoder resolution. |
Zoom_Velocity | 6.0 |
Defines the zoom axis velocity. |
Zoom_Acceleration | 45.0 |
Defines the zoom axis acceleration. |
Zoom_Jerk | 0.0 |
Defines the zoom axis tuning parameter Jerk. |
Zoom_PFactor | 50.0 |
Defines the zoom axis tuning parameter Proportional Gain Factor. |
Zoom_DFactor | 0.0 |
Defines the zoom axis tuning parameter Derivative Gain Factor. |
Zoom_IFactor | 0.0 |
Defines the zoom axis tuning parameter Integral Gain Factor. |
Zoom_ILimit | 0.0 |
Defines the zoom axis tuning parameter Integral Limit. |
Zoom_ContThreshold | 10.0 |
Defines the zoom axis tuning parameter Continuation Threshold. |
Zoom_FollowingLimit | 0.05 |
Defines the zoom axis tuning parameter Following Error Limit. |
Zoom_BacklashCorrection | 500 |
500 |
500 |
500 |
Defines the zoom overshoot distance to counter Backlash (encoder counts) Correction. |
Illumination_SettleTime | 0.1 |
0.1 |
0.1 |
0.1 |
Defines the delay in seconds for the illumination to settle after a change. |
Stage_SettleTime |
0.1 |
0.1 |
0.1 |
Defines the delay in seconds for the stage movement to settle after the software detects a motion. |
Zoom_SettleTime | 0.1 |
0.1 |
0.1 |
0.1 |
Defines the delay in seconds for the optics to settle after a magnification change occurs. |
Pendant_EStopInhibitsAmp |
True |
True |
True |
When set to True, the emergency stop (ESTOP) disables the drive amps. |
Pendant_FineSpeedFactor |
1.0 |
1.0 |
1.0 |
Defines the fine speed as a percentage of the maximum speed. The valid values are 0.0 - 1.0 inclusive (0% - 100%). |
Pendant_X_Reversed |
False |
When set to True, the X-axis pendant motion is reversed. |
Pendant_X_Deadband |
0.2 |
Defines the X-axis pendant motion deadband at the center position in percent. |
Pendant_X_Speed |
1.0 |
Defines the X-axis pendant movement speed. |
Pendant_X_FineSpeed |
0.1 |
Defines the X-axis pendant fine movement speed. |
Pendant_X_Exponential |
0.5 |
Defines the X-axis pendant motion curve. |
Pendant_Y_Reversed |
True |
When set to True, the Y-axis pendant motion is reversed. |
Pendant_Y_Deadband |
0.2 |
Defines the Y-axis pendant motion deadband at the center position in percent. |
Pendant_Y_Speed |
1.0 |
Defines the Y-axis pendant movement speed. |
Pendant_Y_FineSpeed |
0.1 |
Defines the Y-axis pendant fine movement speed. |
Pendant_Y_Exponential |
0.5 |
Defines the Y-axis pendant motion curve. |
Pendant_Z_Reversed |
False |
When set to True, the Z-axis pendant motion is reversed. |
Pendant_Z_Deadband |
0.1 |
Defines the Z-axis pendant motion deadband at the center position in percent. |
Pendant_Z_Speed |
1.0 |
Defines the Z-axis pendant movement speed. |
Pendant_Z_FineSpeed |
0.1 |
Defines the Z-axis pendant fine movement speed. |
Pendant_Z_Exponential |
0.5 |
Defines the Z-axis pendant motion curve. |
Pendant_Q_Reversed |
False |
When set to True, the rotary Q-axis pendant motion is reversed. |
Pendant_Q_Deadband |
0.1 |
Defines the rotary Q-axis pendant motion deadband at the center position in percent. |
Pendant_Q_Speed |
1.0 |
Defines the rotary Q-axis pendant movement speed. |
Pendant_Q_FineSpeed |
0.1 |
Defines the rotary Q-axis pendant fine movement speed. |
Pendant_Q_Exponential |
0.5 |
Defines the rotary Q-axis pendant motion curve. |
Probe_DebounceTime |
0.05 |
0.05 |
0.05 |
Defines the time in seconds that the touch probe must be triggered to detect a hit. |
Probe_DirectionThreshold |
0.01 |
0.01 |
0.01 |
Defines the distance in mm that the touch probe must move to determine the hit vector. |
Probe_DisableIfNotActive |
True |
True |
True |
When set to True, the software disables the touch probe if it is not the active probe. |
Probe_UsePI200 |
False |
False |
False |
When set to True, the software uses the Renishaw PI200/TP200 probe. |
