Step 7 - Take the Third Manual Hit
For the TP20 Probe Changer, you will take the third hit on the front surface and to the left of the last port.
PC-DMIS Message
Please take a hit in the center of the silver front face between ports 5 and 6.
After completing this hit DCC calibration will begin.
Prompt for the third hit on the top to the left side of the last port
Use your machine's jog box to measure the third hit on the top surface to the left of the last port:
Location of third manual hit during TP20 calibration
After it takes this hit on the back of the last port, PC-DMIS drives in DCC mode and takes some additional measurements. It works backwards from the last port toward the first port. It actually measures the same three locations in reverse order in DCC mode. Two additional hits are then taken on either side of port 1.
There is no specific message to indicate that the calibration is complete. The PC-DMIS Execution dialog box closes, and the Calibrate tab reappears.
The next step is to review the calibration results.